Discuss touch leakage testing class 2 items in the Electrical Testing & PAT Testing Forum area at ElectriciansForums.net

1

190ebob

Hello
Can anybody tell about touch leakage testing and when it should be done. We have a seaward europa plus unit that seems to divert to touch testing when we are testing class 2 items. Any info on this would be great.
 
Last edited by a moderator:
Yes the Europa like the Supernova is menu driven. (This is what I assume you are using - the menus I mean).

Touch leakage testing is when (on a piece of class II equipment) mains voltage is applied and the earth lead of the tester traced around suspected vulnerable parts of the equipment casing - to see if there's any substantial - (over 0.10mA) earth leakage current, that can result in a belt if touched.
 
Hello
Can anybody tell about touch leakage testing and when it should be done. We have a seaward europa plus unit that seems to divert to touch testing when we are testing class 2 items. Any info on this would be great.

(see CoP Sect. 15.6)
Touch leakage current testing is not intended as a substitute for insulation testing, but can be used where the Insulation Resistance Test cannot be performed or gives suspect results. Basically, it is a live operating test (so take care!) using supply voltage and the test is performed between internal live parts and the external casing to see if any leakage current can be measured. Current limits:

Class 1: portable or hand-held <0.75mA;
Class 1: heating appliances <0.75 mA or 0.75mA per kW whichever is the greater up to a max of 5mA;
Class 1: other equipment <3.5mA.
Class 2: <0.25mA.
Class 3: <0.5mA.
 
Patman

Yes we are using the europa plus, so do you think for this test to be performed it is down to the user selecting the test?. We have never had this test performed in the past but for some reason it is being used to cover all of our Class 2 items. Thanks for your reply
 

Reply to touch leakage testing class 2 items in the Electrical Testing & PAT Testing Forum area at ElectriciansForums.net

Similar Threads

Hi, I carry out PAT testing on landlord appliances for letting agents in rented accommodation. Occasionally some of the items I cannot access...
Replies
5
Views
712
Hi all, There's 2 single fan ovens in the house my mother recently moved into. Posh elecronic AEG units. It's been sat empty for a year, so the...
Replies
19
Views
1K
Hello, I had an eicr done in 2021and it passed with four C3 items. They recommend next test date to be in 3 years instead of 5 on the certificate...
Replies
15
Views
1K
Hi all, Disclaimer: I'm not an electrician and have very basic knowledge of electrical installation. We've been putting in a new kitchen into an...
Replies
17
Views
2K
Good day. First time poster. We recently had an electrician perform the EICR, as this is a newly purchased property I thought'd I would have the...
Replies
7
Views
692

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock