Discuss megger 1552 problem in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

4

4x4 mark

ive had my megger for just short of a year and today after changing a consumer unit i got a strange reading when testing the rcd trip times.
it will not auto test, ie run through the test sequence by itself, it just displays >50v. if you take each test on its own it will 1/2 test 1x test and ramp test but it will not 5x test just shows same message.
i know you can change test voltages between 25v and 50v but i assume that the megger is at fault as it seems to do this at the output side of the rcd as well as a socket. is this the megger not putting the test voltage out to the circuit, yes the circuit is energised, i also tried changing the batteries but still the same.

cheers mark
 
I wouldnt mess about, Mark.

Get it to the repair shop pronto.

It should still be covered by warranty, assuming you bought new.
 
The Megger shows >50 and stops the test when it suspects that continuing with the test will raise the potential of the earth above 50V. This sometimes happens when the Ze is very high, for instance on a TT system with a unsatisfactory electrode.
Does this tally with what you are testing?
 
tried it on a check box at work and it wouldn't "trip" the check box, so its away at the mo for a health check and calibration. had some ****** auld rubbish off the company i work for took me twice as long to do a test as every time i tried to test a circuit it wanted to put the result in the memory instead of just displaying it.
 
Last edited by a moderator:

Reply to megger 1552 problem in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

Similar Threads

Hi! Some help would be appreciated! I've just moved into new house, wanting to check insulation resestance etc as getting some nuisance...
Replies
2
Views
303
Hi guys & gals, I gave in and bought the new Megger MFT-X1. Got my hands on it Friday night, had a little play with it, used it at work on Monday...
Replies
10
Views
3K
Just thinking about selling MFT's and single test equipment. The idea being I buy stuff from Ebay and sell it on ebay, which I am doing at the...
Replies
0
Views
273
Recently changed a CU for a friend and today when visiting one of the RFC's RCBOs tripped. It was due attention anyway as open ring and no time to...
Replies
13
Views
1K
Please advise what I should test / check next. My usual qualified electrician who did all of the work here is in Ireland for 4 weeks and not...
Replies
45
Views
3K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock