Discuss Metrel insulation testing issue in the UK Electrical Forum area at ElectriciansForums.net

Lestat

-
Reaction score
1
I have a Metrel 17th edition multitester with a potential issue in the insulation testing mode. With no leads connected, I would expect a value of 500-900 MΩ but instead I get something around 50 MΩ.

Am I right to think there's something wrong with it?
 
I would say that there is something wrong.
The reading without leads connected (or even with leads connected but not in contact) should be >1000MΩ.
Initially check that there is no contamination or water tracking across the leads connector point ( and the leads plug), this is the easy hope. If everything is OK there then it is probably something internal which can be much more difficult and costly to resolve.
 
I have a Metrel Mi3101. Testing at 500V my max IR is >1000MOhm, testing at 250the max is >200MOhms, using the easyswitch the max is >30MOhm.
Is the tester new to you? If so I would check the settings.
Can you do an IR test without the leads connected, (not loose but unplugged from the tester)? What result do you get? Could you post a photo of your Metrel screen?
 
My Megger 1730 started exhibiting some strange readings on continuity, with one lead attached to an MET and then as I put the other lead near (not on or connected to) a copper pipe, its started to show a fluctuating low reading. I recall it was doing some other daft things.

After ringing Megger, it went back to them for repair.
 
I have a Metrel 17th edition multitester with a potential issue in the insulation testing mode. With no leads connected, I would expect a value of 500-900 MΩ but instead I get something around 50 MΩ.

Am I right to think there's something wrong with it?

Hi - give Steve Jackson a call in our tech dept on 0113 248 99 66 option 2 - he was Metrel's service manager for many years and he'll be able to diagnose the fault for you over the phone.
 
We have 2 of these testers in work and this is a very common problem, leave it in to get repaired, we were told 1 of ours was repairable but the other was beyond it and it's o my 3 years old . They aren't very reliable overall
 
Maybe I should have mentioned, it's doing 50MΩ with no leads plugged and the back of the unit off, with that plastic socket plate hanging loose. Before I took the back off, it was doing 15MΩ with no leads plugged in. So 50MΩ is actually an improvement and gives me hope the unit is not compromised, but the fault has to be traced still. I'm attaching a picture with the screen.

The way the tester works is, it has 2 wires (L and N) coming off the main board > into the ceramic fuses > then to a 1Ω resistor > and back on the main board. The external test leads connect (T-joint) just after the resistor.

The thing is, if I take the fuses off and test (all test leads unplugged of course), the meter shows 0Ω (continuity). Somehow counter-intuitive, but that's what it does. So in other words, it needs a solid continuous path to show the many MOhms. By this logic, it looks like it might be a dry or cracked solder joint somewhere, rather than a short condition.

Matthewd, do you remember what the problem was with yours?

Test-meter, thanks for the suggestion. I'll give the man a call.

IMG_1416.jpg
 
That is the same one as we had that sadly couldn't be fixed, We had it in 2 different places and the insulation resistance was never correct it was even occasionally detecting a dead short even without lead connected and the Zs values were up the left, reading far too high and when we tested them with the working tester found they were no where near correct. They advised we sent it back to metrel as they had seen it before but for the price the boss opted to spend a little extra on a new fluke one, which I have to say has been great. Sorry for the lack of help I'm not sure what the exact problem with it was
 

Reply to Metrel insulation testing issue in the UK Electrical Forum area at ElectriciansForums.net

Similar Threads

I have been going round in circles with this question. I just cant seem to get an answer remotely close to the options. Continuity of a ring...
Replies
6
Views
1K
Hello All, I have just found out that a family member who is having some Building work done has been advised to insulate above the Kitchen...
Replies
16
Views
745
Insulation test live to earth = 0.7 ohms. This is a long length of buried swa cable (approx 100m) from mains in to submain. Test done with live...
Replies
35
Views
2K
Hi all, We have a limited in house PAT testing requirement and were advised that the TIS700 fitted our needs. The unit works fine for 110V and...
Replies
0
Views
653
I got a TEXT message (i dont know them and i cant find out who they are from the phone number) "Hi speaking from Indeed Are you available to give...
Replies
15
Views
869

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Electrical Forum

Welcome to the Electrical Forum at ElectriciansForums.net. The friendliest electrical forum online. General electrical questions and answers can be found in the electrical forum.
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock