Discuss error4 on fluke 1651b tester in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

N

Nuvo

Having problems with my 1651b fluke mft when measuring my ze when i clip the croc clip on earth and touch probes on phase and neutral it shows error 4 straight away without me pressing test. I've tried this at home too on my board same thing is it time for a new tester or leads. every other test fine so makes me think the leads are fine.
 
Don't have the manual in front of me , but when measuring Zs and putting leads on phase and neutral display shows Error 4 until I put the earth lead on. If I stil get Error 4 then I start suspecting there is no earth!

If its your leads I would expect the zero'ing values to be changing. If you've had the tester a longeish time you should know roundabout what they usually are.
 

Reply to error4 on fluke 1651b tester in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

Similar Threads

So, I was messing around yesterday with some old testers I have laying around to see if they still work I found my dads old Robin MFT and noticed...
Replies
8
Views
552
Hi All New to this forum, have read the posts on here from google but only recently signed up. I'm having some issues and some input would be...
Replies
13
Views
2K
Hi all, We have a limited in house PAT testing requirement and were advised that the TIS700 fitted our needs. The unit works fine for 110V and...
Replies
0
Views
708
Having a challenge understanding the concept behind why when you measure across line and neutral (live circuit) using voltage indicators we have...
Replies
25
Views
2K
Fluke MultiFunction Tester 1662 TNS Earthing system. Switchboard - 415v - 3 Phase + Neutral + Earth. MCB 1000A 3 individual Earths (185mm)...
Replies
22
Views
3K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock