Discuss Megger 1552 (Hi current test not tripping 30ma rcd) in the UK Electrical Forum area at ElectriciansForums.net

Reaction score
1
Hi guys, I own 2 megger mft 1552 testers.

I recently found that loop readings on both testers give me high Zs readings on final circuits protected by 30ma rcds on low current.
If I select the hi current setting I get a reading I expect to achieve, but the hi current test never trips the rcd.

Both testers have been calibrated in the last 3 months

Any advice?

Many thanks David
 
What make are the RCD’s as I had the same issue with Hager RCBO’s a few years ago. They all tested fine whilst doing the RCD trip times.
 
the Rcds are MK mainly but it’s recently happend on a CPN rcd aswell. It seems to be more on the recent metal clad dbs, I don’t know whether they have altered the rcds in anyway
 
A workmate had the same model and was getting silly loop readings. Bought a new set of leads thinking it was that, but they made no difference, work sent it away for repair and calibration but was still the same when it came back so it's no longer used. The test readings have never really been very repeatable with it.
 
It’s on both tn-s and tn-c-s. I’m yet to use it on a tt system.
I have spoken to megger technical. They said to use the hi current reading and just to do a ramp test to ensure the rcds are tripping at an acceptable Ma. Just seems bizarre it’s with both my 1552 testers
It's good that you have 2 testers as that kind of signifies that the problem is with the 'type' of tester rather than perhaps an actual fault with the tester or a fault in the installation. It would be even better if you could try a different type of MFT on the same RCD's then you could narrow the problem down depending on if the different MFT managed to trip the RCD on high current.
 
It's good that you have 2 testers as that kind of signifies that the problem is with the 'type' of tester rather than perhaps an actual fault with the tester or a fault in the installation. It would be even better if you could try a different type of MFT on the same RCD's then you could narrow the problem down depending on if the different MFT managed to trip the RCD on high current.

I asked a guy that works with me to get his tester out aswell, which was only a week old. His tester was working accurately.
We done a loop test using the megger and his di-log on either side of the rcd. The megger recorded .50ohm difference, where as his tester recorded no difference at all.
 

Reply to Megger 1552 (Hi current test not tripping 30ma rcd) in the UK Electrical Forum area at ElectriciansForums.net

Similar Threads

What should you do if a client has a CU that is a split board protected by 2 * 63Amp 30mA double pole RCCDs, you took Zs and tested both RCCDs by...
Replies
22
Views
2K
Hi all, Been a while since I have been on here. I have been on an apprenticeship the last 3 years training in the BMS world. Taking that into...
Replies
7
Views
320
Keeping this brief.....................New dwelling.....TT system with 100mA S type rcd protected supply to local board with 30mA RCD. Freelance...
Replies
21
Views
3K
Called out to fault on RCD tripping maybe twice in a month for sometime. Did all the tests & found RCD was faulty, Refitted a new Rcd Type A which...
Replies
2
Views
785
I just finished a garden room supply. The customer already had a CPN garage consumer unit that the last sparks left there, brand new in box, and...
Replies
2
Views
877

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Electrical Forum

Welcome to the Electrical Forum at ElectriciansForums.net. The friendliest electrical forum online. General electrical questions and answers can be found in the electrical forum.
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock