157
Here's a tip for anyone still using an old MFT1700 series tester. Mine dates from 2012.

I found about three years in - by 2015 in this case, that the zeroing of the leads on my MFT1720 was becoming increasingly random and the zero would wander about by a fraction of an ohm which became irritating and time consuming. Others talk of replacing leads -(Kewtech are good, and this is certainly worth doing, but it's's not the only problem). At the end of the day, a copper lead is just that. there's no magic. It's the contacts which make the difference.

Cleaning the external leads and instrument terminals with Isopropyl alcohol and cotton buds helps. - But the big win on these ageing instruments is the contact between the terminals "pod" and the main board.

So if you are interested, here's what you do.
Turn the instrument over, remove the batteries, contacts towards you. Remove the two screws adjacent to the fuses.
In between the two fuses you see a rectangular hole and a square one. The rounded plastic boss you can see down the rectangular opening is a robust plastic part of the banana-terminals"Pod". The Pod is the oval-shaped moulding which holds all contacts facing your tummy now. The rear of the terminals are socketed into the main board with male gold-plated spring pins.

Having removed those two screws, nothing except friction holds the "pod" into the case.

So, get a wide blade flat screwdriver and insert it into the rectangular hole, then prise it to push the grey "boss" (part of the oval contacts pod) out of the front of the instrument. You may find first time it's really tight, and one end of the pod wants to release firsts. Hold it straight and lever it out. First time will possibly need quite a bit of force.

The moulding comes right out, and bingo, the source of your problem. You will see spring gold plated pins, which in the factory are lubricated with some Contra-Lube (or similar grease).
These gold contacts can get a little grubby, and a small build-up will add resistance and cause your never ending zeroing problems.
Clean them a little very gently, - if you like with a little isopropyl alcohol. Add a dab of Contra-Lube and push the pod back in.

Bingo, prefect zeroing. I find annual of six-monthy maintenance helps, but I will always do it if I feel the zeroing is beginning to wander a little.
 
still does not get round the problem of being unable to clean the ends of the leads as they are now secreted within shrouds and inaccessible.
 

Reply to MFT1700 series wandering zero problems fixed. in the Electrical Tools and Products area at ElectriciansForums.net

Similar Threads

S
Hi, here's the instructions for my home made wander lead. A lot cheaper than the retail versions and only takes a few minutes to put together...
Replies
7
Views
9K
BluSparx
B
T
Hi all, great site and sorry for the long post - I'm after some advice as I don't have total confidence in the 4 sparkies I've currently used with...
Replies
30
Views
6K

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock