Search for tools and product advice,

Discuss Digital Mutimeter Calibration in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

S

slackboy grim

We have just been Audited as a company and one of the things they picked up on was that we have no process in place to check calibration monthly on the multimeter the engineers use.
Does anybody know of a check box on the market that will check, Voltage, Resitance and current?
 
If you're company is of any size, then it would be prudent to purchase a decent full size proving/checkbox test facility. For example, such as the Megger MTB7671/2 or Transmille 2080 and set up a regular in-house documented testing process of all company electrical test equipment....
 
What's wrong with a multimeter?
Try fault finding on electronic equipment to component level with an MFT!
Nothing wrong with owning a decent one, I know I have a laugh at times regarding them but I own a good one and it is a fantastic bit of kit, very useful, I use it quite a lot actually, so I understand your post.
 

Reply to Digital Mutimeter Calibration in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

Similar Threads

Just thinking about selling MFT's and single test equipment. The idea being I buy stuff from Ebay and sell it on ebay, which I am doing at the...
Replies
0
Views
325
I've just picked one of these up on a famous auction site. The voltages, low and high resistance ranges all appear to be working fine but it's not...
Replies
7
Views
303
Hi! Some help would be appreciated! I've just moved into new house, wanting to check insulation resestance etc as getting some nuisance...
Replies
2
Views
374
I am working at a site with around 100 of these meters installed on various loads. Doing an audit, we found that most of the meters were recording...
Replies
2
Views
920
Hi all. We have a 120v 20 amp circuit to an accessory dwelling with a run of about 75' of direct bury #12. The breaker keeps tripping even when...
Replies
3
Views
340

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock