Discuss Fluke 1652B "FUSE" fault? Urgent! in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

Reaction score
7
Hello

I've got a Fluke 1621652B which I've been using without fault for the last couple of years.


Was doing some fault finding on a job this morning, random RCD tripping issue, when I was using the ramp test function to check the installed RCD.

Tried first test with all load disconnected to test RCD and result was fine. Reconnected all outgoing circuits and turned basics on (fridge, etc...), tried to ramp test from a local socket, RCD tripped instantly, Fluke had nothing on the screen (no result or anything).

I then tried to use the fluke again to which it came up with 'FUSE' on the screen. I thought this was strange however attempted to replace the internal fuse.

I didn't have any of the F1A fuses handy however had an F10A fuse of same size which wasn't blown (verified with voltstick continuity test) so tried this in order to verify there was no problem with tester (original fuse had in fact blown)

Unfortunately the tester still comes up with the FUSE fault, even with a working fuse fitted. There is a loud click from within the tester before the FUSE fault comes up.

Whats the prognosis? Is it knackered? Does anyone have any ideas of costings to repair? I've got solid EICRs booked in for the next few days and could really do without this hassle, should have had a spare!
 
Hello

I've got a Fluke 1621652B which I've been using without fault for the last couple of years.


Was doing some fault finding on a job this morning, random RCD tripping issue, when I was using the ramp test function to check the installed RCD.

Tried first test with all load disconnected to test RCD and result was fine. Reconnected all outgoing circuits and turned basics on (fridge, etc...), tried to ramp test from a local socket, RCD tripped instantly, Fluke had nothing on the screen (no result or anything).

I then tried to use the fluke again to which it came up with 'FUSE' on the screen. I thought this was strange however attempted to replace the internal fuse.

I didn't have any of the F1A fuses handy however had an F10A fuse of same size which wasn't blown (verified with voltstick continuity test) so tried this in order to verify there was no problem with tester (original fuse had in fact blown)

Unfortunately the tester still comes up with the FUSE fault, even with a working fuse fitted. There is a loud click from within the tester before the FUSE fault comes up.

Whats the prognosis? Is it knackered? Does anyone have any ideas of costings to repair? I've got solid EICRs booked in for the next few days and could really do without this hassle, should have had a spare!


Contact FLUKE and see what they say, better still take it to your local wholesaler and get them to send it off for repairs
 
I actually did call FLUKE who's engineer was supposed to be calling back but the call never came

Even better, took it to a local company who actually do the repairs themselves rather than a wholesaler, it turned out to be loose contact springs on the fuse, apparently if there is a high resistance joint the tester sees it as faulty

The guy even threw in a few spare F1A fuses

If anyone is from the Yorkshire area, PJO industrial in Barnsley are a good company to do business with, saved my skin today!
 
Last edited:

Reply to Fluke 1652B "FUSE" fault? Urgent! in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

Similar Threads

Hi Guys, I'm having trouble diagnosing a fault on a customers downstairs socket ring, which only ever trips when the washing machine is used. The...
Replies
22
Views
3K
Hi All New to this forum, have read the posts on here from google but only recently signed up. I'm having some issues and some input would be...
Replies
13
Views
1K
Please advise what I should test / check next. My usual qualified electrician who did all of the work here is in Ireland for 4 weeks and not...
Replies
45
Views
3K
Good evening, I have recently moved into a new home and I am having problems with the MCB/RCD tripping (Mem M6 Type 3 - 30mA). It intermittently...
Replies
8
Views
1K
Hi all, There's 2 single fan ovens in the house my mother recently moved into. Posh elecronic AEG units. It's been sat empty for a year, so the...
Replies
19
Views
1K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock