Discuss Megger PAT2 replacement probe in the Electrical Testing & PAT Testing Forum area at ElectriciansForums.net

Reaction score
15
Hi all, I'm new to both this forum and PAT testing. I've been given a Megger PAT2 to use but it's missing its flash testing probe, so I can't use the tester for Class 2 insulation tests. I'm not hopeful of getting a spare from Megger themselves so can anyone advise on a replacement? I'm afraid I don't know what the fitting is. Thanks.
 
Welcome to the forum.

The PAT2 tester is now rather old and it may be difficult to get spares for it, but there do seem to be quite a few PAT2 on ebay and the like and it may be possible to get an entire replacement unit for less than the cost of a new probe.

However I am not sure what test probe you are missing as a flash test is one that is undertaken for type testing of a piece of equipment under initial construction and is very rarely used to test existing equipment and certainly not in a PAT environment.

Class II insulation tests are undertaken using line and neutral joined together and using a probe at points of potential weakness of the insulation or points where conductive material may accumulate.
 
Welcome to the forum.

The PAT2 tester is now rather old and it may be difficult to get spares for it, but there do seem to be quite a few PAT2 on ebay and the like and it may be possible to get an entire replacement unit for less than the cost of a new probe.

However I am not sure what test probe you are missing as a flash test is one that is undertaken for type testing of a piece of equipment under initial construction and is very rarely used to test existing equipment and certainly not in a PAT environment.

Class II insulation tests are undertaken using line and neutral joined together and using a probe at points of potential weakness of the insulation or points where conductive material may accumulate.
Point taken about flash testing but, if I'm understanding the operation of the PAT2 correctly, the same probe doubles as the insulation testing probe.
 
Point taken about flash testing but, if I'm understanding the operation of the PAT2 correctly, the same probe doubles as the insulation testing probe.

That's correct (see here, page 11: http://www.biddlemegger.com/biddle-ug/PAT2_UG.pdf)
Same as the old Seaward PAT1000S & PAT1000IT units: black croc-clip lead for Earth Bond, red spring-loaded probe for Class II Insulation and Flash tests.

Later models by most (if not all) manufacturers ditched this arrangement, and had the clip lead for both Earth Bond and Class II Insulation, while the red probe only does Class II Flash.
 

Reply to Megger PAT2 replacement probe in the Electrical Testing & PAT Testing Forum area at ElectriciansForums.net

Similar Threads

Hi all, We have a limited in house PAT testing requirement and were advised that the TIS700 fitted our needs. The unit works fine for 110V and...
Replies
0
Views
674
  • Article
Electrical2Go - Megger MFT Deals! Upgrade your toolkit with our exclusive offer! Purchase any of the Megger advanced multifunction testers –...
Replies
0
Views
599
Hello, I am currently learning and have come across PAT testing an old chandelier. On modern lamps I would use the plug and probe on the...
Replies
1
Views
2K
Hi! Some help would be appreciated! I've just moved into new house, wanting to check insulation resestance etc as getting some nuisance...
Replies
2
Views
318
So bought a second hand Megger BM233 to practice with as I'm a college student.. So I was just using it on a loose 2.5 cable. Testing on the...
Replies
11
Views
983

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock