Discuss Megger 1730 MFT. Not tripping on any RCD tests on known good circuit. in the UK Electrical Forum area at ElectriciansForums.net

Marti

-
Arms
Reaction score
244
G'd Evening Lads an' Lasses,
I've got a Megger 1730 MFT; lovely bit of kit normally. Ex-demo - not had a hard life and was calibrated about eight weeks ago. I normally run it over a Cal-Card and a captive test RCD about once a month and this time it failed to trip the 30ma RCD on 1x, 5x and ramp. Tested all the RCD's in the house - same result. Reset everything to default - same result.

Has anyone seen anything like this? Really appreciate your thoughts. just started out on my own so the timing is ERR...unfortunate if it's terminal.
 
Was it definitely set for a 30mA test and not 300mA.
 
G'd Evening Lads an' Lasses,
I've got a Megger 1730 MFT; lovely bit of kit normally. Ex-demo - not had a hard life and was calibrated about eight weeks ago. I normally run it over a Cal-Card and a captive test RCD about once a month and this time it failed to trip the 30ma RCD on 1x, 5x and ramp. Tested all the RCD's in the house - same result. Reset everything to default - same result.

Has anyone seen anything like this? Really appreciate your thoughts. just started out on my own so the timing is ERR...unfortunate if it's terminal.
Send it to these guys Services - https://www.alpha-electronics.com/services/most of their team are ex Megger as only down the road.
 
Surely if set for 300mA test on a 30mA RCD it would be more likelyt o trip it?
True. That’s me jumping in again without thinking properly.
 
Oh, sorry should have mentioned the obvious. Tested the same with the Megger leads in a breakout and the 1363 plug lead. Both could be faulty but unlikely. Also, when my test RCD failed I used it on all the house circuits but the timer just keeps counting up and eventually display "greater than 40ms". My workshop is TT with a Ze of 75-80 ohms. The house is standard TNC-S. All tested about six weeks ago.

Thanks for all the input people
 
Had a similar issue when one of my 1720 models went for calibration I’m not sure what they do to it but it worked for one Zs test and then when I switched to do an RCD test I got nada being a curious guy I thought I’d have a look inside I wouldn’t recommend it but that particular meter owed me nothing anyway a bit of solder has melted and a wire jumper had become detached resoldered and checked it - it worked fine so sent it back for recalibration and it was all good if anything the tolerance was slightly better than when compared to the previous cal sheet. Opening the thing was a bugger though that’s why I wouldn’t recommend it it could be very very easy to ruin it...
 
Was it definitely set for a 30mA test and not 300mA.
You may have been wrong , but Operator error (Brain fart) is always possible , the more sure you are of your self .
- Once you remember the wrong way .. Or have mft remember it for you
( miss some obscure setting again) again .. a .. a .. !
 
Are you sure you've put the leads into the live and earth into the megger, not the live and neutral. I'm sure you wont have done as it sounds like you've had it for ages but worth saying.
I've only had my megger a short while and I did that by accident and thought.... arghh the RCD's not tripping!
 
Update on the poorly Megger.
Took it into Barking and Dagenham College (did my C&G there) They tested my MFT and leads on a college rig with a known good 1720. Definitely a faulty unit. The only question is whether I caused this switching between functions whilst powered? Don't know but probably worth bearing in mind. Thanks for all the input guys.
 
G'd Evening Lads an' Lasses,
I've got a Megger 1730 MFT; lovely bit of kit normally. Ex-demo - not had a hard life and was calibrated about eight weeks ago. I normally run it over a Cal-Card and a captive test RCD about once a month and this time it failed to trip the 30ma RCD on 1x, 5x and ramp. Tested all the RCD's in the house - same result. Reset everything to default - same result.

Has anyone seen anything like this? Really appreciate your thoughts. just started out on my own so the timing is ERR...unfortunate if it's terminal.
Are you testing the RCD at the RCD OR via a socket outlet?
 
Hi Pete,
Tried both. The meter skips from test straight into the timer. Our 1720 comparison unit on 1/2 delta tested for 3 - 4 seconds then counted. Mine went straight to count. The 1720 then tripped on 1x and 5x delta. Mine just counted so it's going to Alpha.
 
Hi Pete,
Tried both. The meter skips from test straight into the timer. Our 1720 comparison unit on 1/2 delta tested for 3 - 4 seconds then counted. Mine went straight to count. The 1720 then tripped on 1x and 5x delta. Mine just counted so it's going to Alpha.
OK thanks, are the outgoing circuits still energised?
 
Update on the poorly Megger.
Took it into Barking and Dagenham College (did my C&G there) They tested my MFT and leads on a college rig with a known good 1720. Definitely a faulty unit. The only question is whether I caused this switching between functions whilst powered? Don't know but probably worth bearing in mind. Thanks for all the input guys.


Hi Marti,

Have you tried dropping it in to these guys.......

Test Instrument Calibration - Electronic Instrument Calibration Repair - Essex, UK - http://www.capitalcalibration.co.uk/

They are pretty helpful and calibrate and repair our testers all the time.
 
Hi Pete,
Tried both. The meter skips from test straight into the timer. Our 1720 comparison unit on 1/2 delta tested for 3 - 4 seconds then counted. Mine went straight to count. The 1720 then tripped on 1x and 5x delta. Mine just counted so it's going to Alpha.
 
Not familiar with the tester, but what do you mean by just counted?

It's not set for time delayed RCD's or the like?
 

Reply to Megger 1730 MFT. Not tripping on any RCD tests on known good circuit. in the UK Electrical Forum area at ElectriciansForums.net

Similar Threads

Please advise what I should test / check next. My usual qualified electrician who did all of the work here is in Ireland for 4 weeks and not...
Replies
45
Views
3K
Hi guys & gals, I gave in and bought the new Megger MFT-X1. Got my hands on it Friday night, had a little play with it, used it at work on Monday...
Replies
10
Views
3K
I have just popped out to a frustrated customer who has had intermittent tripping of a 30mA RCD for years, on average 5 - 10 x p/month. It used to...
Replies
31
Views
4K
Hi, quick question before I start ripping a customers Consumer Board apart. Called out to a simple job of installing electric supply to a summer...
Replies
6
Views
5K
R
Testing 2 RCD's in a starbreaker dual board today. MCB's off and testing to the busbar both RCD's @5x, both RCD's gave exactly 4.8ms. Far too...
Replies
2
Views
1K
radiohead
R

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock